I don't work with Agilents, so I can't help with the technical side.
Why do you not want air in the needle ?
Peter
Thank you, i want to make sure that the the amount injection is correct.
...
if I set 1 ul > syringe pump 3 times and i see 2ul level in barrel
if i set 2ul > the syringe pumps 4 times and I see 3 ul level in barrel
Thank you
If you mean "pumps" = syringe wash than number of pumps has nothing to do with injection volume and should not change when you change injection volume.
I am sorry for my poor English, I mean that If i set to inject 1ul in the chemistation - for the front injector (from desktop computer) then I check the GC system, by press to the Front Injector (in the method storage and automation) I see in the small screen in GC system appear:
Front injector
Injection Vol 2
#Sample pums 3
...
Here I mean that the GC system automatic convert to 2 ul, even I set the front injector as 1ul.
Similarly, when I set to inject 0.2ul, the GC will move to 1 il, 1 to 2, 2 to 3, 3 to 4 and 4 to 5.
If i changed the syringe volumes to 5 ul, the similar thing was also happened,
0.1 to 1,
0.5 to 2
1 to 3,
1.5 to 4
and 2 to 5
So I am wondering if how the volumne in GC system can be understand: it is the volumne that actual amount of sample to be injected, or the is number to indicated the level to be injected from chemistation setting.
Moreover, I also saw that the niddle inside synringe moved according to the volumne setting from GC system. That what I mean in the previous post, if I set inject 1 ul from chemistation > GCsystem read as inject 2ul > The neddle move up two times when taking the sample (so I observed that the head of needle is at 2 ul in barrell).
I hope it is clear to you. Thank you