We have recently purchased a Shimadzu LCMS-2020 (ESI) single quad system, and I attempted to run a method that I use on a HILIC column on a HPLC-ELSD system (that works beautifully) to try and identify a particular peak, but also to get the mass of the main peak. (By NMR this poly sulfated sample is 90% pure).
The mobile phase is: A:5mM NH4 Formate; B:5mM NH4 Formate in 95%Acetonitrile and 5%Water
Gradient is 3%A:97%B to 15%A:85%B in 14 mins, hold for 5 mins and then back down again.
Chromatographywise, the peaks come out about the same time as they did on the HPLC-ELSD system (as both systems have a splitter). However, some peaks are negative (in either negative or both neg and pos scan modes). And the main peak looks like a saw. That is the tail is jagged. And the mass I obtained across this peak varied, none of them being the correct mass of the poly sulfate.
My question is: These poly sulfated compounds are very unstable. And so I am thinking the sulfates are falling off because my MS conditions are too harsh. The mass spec is tuned to the tuning solution, but if the tuning solution is nothing like the sample then perhaps I cannot use the MS default values, and I have to optimise the conditions. This is what I have no idea about. What parameters should I look at changing first to try and be a bit more gentle? That is what parameters do Mass spec'ers change when optimizing?
I would be so grateful for any help, as this MS detector works beautifully with other compounds (non sulfated) I have tried on it, and used the default conditions.