- Posts: 4
- Joined: Fri Apr 18, 2014 4:35 pm
I have cleaned the source, replaced the traps, installed a new column, and cleaned the transfer line and inlet with methanol but am still having the issue. I recently changed the desorb time from 1 minute to 4 minutes to be compliant with 624, and I suspect that might be contributing to the problem. The baseline is higher than before - Could this be due to too much water entering the system?
Here are the parameters I am using:
Agilent 6890 GC & 5973 MSD w/ helium
Vocarb K trap on Encon Evolution
Purge: 11 minutes @ 30C & 40 mL/min, Dry Purge: 1 min & 40 mL/min, Desorb preheat: 250 C, Desorb: 4 min @ 260C, Bake: 7 min @ 260C
Column: DB-624 - 0.18 mm ID, 20 m length, 1 um film
Temperature program: 45C initial for 2 min; rate 15 to 250C
Split ratio 50:1
Flow: 0.8 mL/min
Split flow: 39.6 mL/min
Inlet: 150C & 16.8 psi
MSD Quad: 150 C
MSD Source: 230 C
If anybody has any ideas about this I'd appreciate the help.