Does anyone know how to switch the lamps in a ProStar 335 detector off remotely via Star Chromatography workstation? We sometimes have runs that finish in the early evening which means the lamps are left on for several hours doing nothing.
-
- Posts: 56
- Joined: Fri Aug 19, 2011 4:49 am
Hi,
You can set up a sequence list (file-> new squence list). The first line is inject (for action) and enter method name and sample list name. On the next line of the sequence, the pull down menu under "action" gives you a series of options including lamp off and pump off.
If you wish, you can drop the flow rate on the pump prior to turning the lamp off by having the second line of the sequence as "inject", then a different method (with a low flow rate), and a separate sample list (personally I have a vial of solvent in the final position on the autosampler and call it dummy so it never gets confused with real sample lists).
In terms of lamp life, my understanding is two things are at play: how long the lamp is used and how frequently you turn it on and off. It was suggested to me that sometimes you're better to leave the lamp running if you're going to be using the instrument the next day.
You can set up a sequence list (file-> new squence list). The first line is inject (for action) and enter method name and sample list name. On the next line of the sequence, the pull down menu under "action" gives you a series of options including lamp off and pump off.
If you wish, you can drop the flow rate on the pump prior to turning the lamp off by having the second line of the sequence as "inject", then a different method (with a low flow rate), and a separate sample list (personally I have a vial of solvent in the final position on the autosampler and call it dummy so it never gets confused with real sample lists).
In terms of lamp life, my understanding is two things are at play: how long the lamp is used and how frequently you turn it on and off. It was suggested to me that sometimes you're better to leave the lamp running if you're going to be using the instrument the next day.
2 posts
Page 1 of 1
Who is online
In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 11 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 11 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science