I am looking for a solution to my problem. After maybe two months I wanted continuing my GC measurements on my Agilent system and every method I load can't start anymore and gives out the following message:
"-222,"({SA00} :MSC:PARM 15:24:12 10/28/2015
R MASG,-999999999999999970000000000000000000000000
R 000000000000000000000): Data out of range" "
I am not sure what I can do there and can't seem to find any proper error codes regarding this problem. The error happens just before the run is supposed to start.
Does anybody of you have an idea how to fix this problem? Of course I checked all connection, restarted computer and GC/MS and reconnected network cables etc. Before all the saved methods were running perfectly. Now not a single method I can start.
I am working with the "GC/MSD Systems - 5975C Series" and using the MSD ChemStation Software.
I hope you can help me and wish you all a nice week. Thank you very much,
Chris
EDIT: Problem solved by deleting all connections to the computer and reconfiguring with "Agilent MSD Configuration" ...So simple