Low recoveries for EPA Method 508
Discussions about GC and other "gas phase" separation techniques.
3 posts
Page 1 of 1
I'm having recovery issues with EPA Method 508. We use the mechanical method for the liquid-liquid. None of the GC parameters have changed from the last time things were run but we have always been on the lower end of passing a PT. Currently we are using Teflon TCLP bottles for the tumbling and Teflon Sep. funnels. If anyone has ANY suggestions on improving either the extraction end or the GC method please let me know. Thanks
-
- Posts: 2
- Joined: Sun Dec 20, 2015 11:38 pm
Hello - I had a quick look at EPA 508 - it looks similar to the 8081 and 8082 methods.
I assume you are extracting with DCM and then solvent exchanging into hexane for ECD analysis or just running in DCM on an MS???
At any rate, my last lab that I worked with experimented with a tumbler style mechanical liquid-liquid extraction and found our recoveries dropped significantly. The tumbler just wasn't rigorous enough when it came to performing the extractions - so we continued to extract by hand. It took longer but our results were much better. You may want to try a few by hand to see if this could be the case for your lab.
Also, if you concentrate your extract too quickly or at too high a temperature, you will lose your analytes. Try dropping the concentration temp to 35-40 Celsius.
What type of concentrator do you use?
I assume you are extracting with DCM and then solvent exchanging into hexane for ECD analysis or just running in DCM on an MS???
At any rate, my last lab that I worked with experimented with a tumbler style mechanical liquid-liquid extraction and found our recoveries dropped significantly. The tumbler just wasn't rigorous enough when it came to performing the extractions - so we continued to extract by hand. It took longer but our results were much better. You may want to try a few by hand to see if this could be the case for your lab.
Also, if you concentrate your extract too quickly or at too high a temperature, you will lose your analytes. Try dropping the concentration temp to 35-40 Celsius.
What type of concentrator do you use?
-
- Posts: 4
- Joined: Wed Oct 28, 2015 4:56 pm
We are exchanging into MTBE and running on an ECD. We use a horizon xcelvap for our concentrations. What were you using at the previous lab?
3 posts
Page 1 of 1
Who is online
In total there are 61 users online :: 0 registered, 0 hidden and 61 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 61 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 61 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science