Check that you GC runtime and sample equilibration time do not lead to the autosampler trying to inject at the same time as it should be loading a sample.
There might also be a simple malfunction in the GC that stopped it being ready. Is this kind of stoppage a regualr event ?
Peter
Hi all,
I'm working with a TD-GC-MS system.. Agilent 7820A + 5975c + Unity 2 Thermal desorption Unit. In fact this last component acts as a injector device for the GC. After using successfully the system for several months, now in the Chemstation appears a similar error than the initially posted. In fact it does not allow me to proceed with my runs.
-310, "({AHI} ): 5201 Timeout waiting for APG RCI ready" 5201, "({AIH} ): Timeout reading APG. Controller interrupt cause".
Evidently in this case I don´t use an autosampler and the injection is replaced by the rapid desorption of my analytes directly into the column. The TD software is somehow link to the chemstation so once the desorption is done, the method can start running. Even when the timer starts, the MSD never does nothing. the filament keeps turned off and consequently there is no data acquisition.
Have you ever deal with this kind of problem? any suggestion would be very helpful.....
Thanks,
Oscar