Yeah, your detector signal indicates shorting out, time to take stuff apart and clean, or at least try hotter FID temperature overnight to see if that dries it out.
And you're complaining of WATER in your house air ??? At our old lab, after about 25 years the air compressor went down, and to "save money" over our objections they purchased an oil-lubed replacement, swearing that the maintenance and safeguards would not let any oil into the house air.
Guess what ? That's right, the whole building air supply became contaminated with oil; so we used glass-sided traps with cotton in them to absorb the oil, and to give us a visual determination of effectiveness. We were lucky, too, as our GC air purifier was mounted up high on the wall, and gravity helped oil from getting into that.
We also had a 5890 GC shipped to us from a warehouse at manufacturing that was saturated with oil, someone had hooked it up to an oil-lubed air supply there. There was liquid oil all through the FID and its controls, and the insulation surrounding the column oven was also saturated, we cut away as much as we could, replaced with fiberglass tape. We could take the GC up to 190 C without smell or smoking after that, so we dedicated it to a single lower-temperature analysis.