Peak present but no quant...
Posted: Wed Oct 14, 2015 2:24 pm
by wenleonard
I have a compound that eludes at 8.541, and got a nice tall normal peak there, however the quant ions for the compound were not present so the compound was not determined in my report even though the peak suggests otherwise. What gives?
Re: Peak present but no quant...
Posted: Wed Oct 14, 2015 2:41 pm
by Peter Apps
Do you honestly expect sensible answers when you give no information at all about what it is that you are trying to analyse, or how you are doing it ?
The more you don't tell us, the more we can't help you.
Peter
Re: Peak present but no quant...
Posted: Thu Oct 15, 2015 7:26 pm
by James_Ball
I have a compound that eludes at 8.541, and got a nice tall normal peak there, however the quant ions for the compound were not present so the compound was not determined in my report even though the peak suggests otherwise. What gives?
As Obiwan would say, "These aren't the peaks you are looking for."
If the quant ions are not present, the possibly the peaks you see in the total ion chromatogram are different analytes then the ones are are trying to quantify. Does the spectra of the peak match the spectra of your target analyte?
Is this GC/MS or LC/MS and what brand instrument are you using and what software version.
Re: Peak present but no quant...
Posted: Mon Oct 19, 2015 10:32 am
by carlo.annaratone
Don't you read? The compounds eludes. It's the famous elusive compound, no wonder even Leonard knows nothing about it.
Re: Peak present but no quant...
Posted: Mon Oct 19, 2015 10:37 am