6890 FID Air Flow won't Turn Off
Posted: Tue Oct 13, 2015 5:22 pm
FID Air flow will not reduce to setpoint. Front FID will light, but the rear FID blows out the flame as soon as the air flowrate increases. With all flows off, the Air flow on rear FID is 120 mL/min. On front inlet it is stuck at 22 mL/min. I have confirmed these flows with a flowmeter. Air flow will drop to 0 if air is blocked with external valve in the line.
Air line pressure is 80-100psi. I brokedown the FID and inspected it, very clean, jet is not plugged, and column installed properly. Column flows are as they should be through the FID.
So, is there an internal FID flow controller or valve that could be stuck open or allowing air to get by & can I fix it or will I have to order the new part?
Thanks,