LabSolutions/Shimadzu LCMS Software Issue...

Posted: Wed Oct 07, 2015 7:53 pm
by am9000
Hi forum,

I'm having an issue with the Lab Solutions software in that even though I am not overloading the detector, when I look at the chromatogram, I am seeing my peak go flat (at the top of the peak) for about 20 seconds. I get the feeling I'm wrestling with a display setting I don't know how to change, especially since I've run the exact same sample with another method (same running parameters) back-to-back and don't have this problem.

Any help?

Thank you!

Re: LabSolutions/Shimadzu LCMS Software Issue...

Posted: Fri Oct 09, 2015 12:24 am
by cene
Just to check the obvious... are you viewing the chromatogram at an appropriate zoom level? Right click on the chromatogram and select "Initialize Zoom". Alternatively you can right click and select "Display Setting". Then tick the "Full chromatogram" checkbox and enter 100% for the area setting. That shows you the entire chromatogram regardless of how many peak groups or data acquisition periods there were in the run.

Re: LabSolutions/Shimadzu LCMS Software Issue...

Posted: Fri Oct 09, 2015 2:11 pm
by dr_bahram1977
What is your detector? LCMS? or PDA?
Check your offline data and see if error persists there or not?