Calibration Table for PCB

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi, I am confused on PCb analysis
Both PCB 30 and PCB 209 are internal standart for PCB analysis.
I have prepared solutions of 0 to 100 ppb(5 level) of PCB mix calibration solution that already have PCB 209
I have add PCB 30 to all five levels so that final consantration of PCB 30 is 25 ppb in all level.

How do I create a calibration curve on ınstrument? I mean I know how to do it. But I don't understand the logic here. On the instrument there are two old method, one of them PCB30 is selected as ınternal standart, in the other PCB209 is selected as internal standart.

Which one is correct or why I don't choose both of them as internal. There is also a option of reference next to internal. Is it something I have to choose.
Either one of them could be correct, depending on what you are doing. If you want to use both as internal standards then you split the target PCBs so that the early eluters (up through tetra chloro) use pcb 30 and everything later uses 209.

Note that if your sample has Aroclor 1268 in it there will be extra 209.
2 posts Page 1 of 1

Who is online

In total there are 129 users online :: 2 registered, 0 hidden and 127 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot], Semrush [Bot] and 127 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry