ICH Q2 (R1) validation question
Posted: Thu Oct 01, 2015 9:22 pm
This is a general question regarding accuracy assessment during method validation.
ICH states in Section 4.2 of the Q2 (R1) guideline that, "accuracy should be assessed on samples spiked with known amounts of impurities". I define "should" in this statement to mean "obligation or duty". A colleague of mine defines it to mean "probable/probably", and has used this position to validate impurity methods using solutions containing only the impurities - no drug substance is present. The rationale is the impurities are present in significant amounts and it is not possible to achieve the concentrations required.
I contend this is not validation as it is performed outside of the actual sample matrix, that the spiked test solutions should be corrected for the impurity amount determined using control solutions.
I can not find any clarification in the ICH documentation. How do others interpret this guideline? Any supporting documentation. What do you do when impurities are present?
I know the FDA's recent Method validation guidance document of July 2015 clearly states in lines 66-70 that "In general, FDA’s guidance documents do not establish legally enforceable responsibilities. Instead, guidances describe the Agency’s current thinking on a topic and should be viewed only as recommendations, unless specific regulatory or statutory requirements are cited. The use of the word should in Agency guidances means that something is suggested or recommended, but not required.
Robotjock
Robotjock