HPLC signal at -500 mAU

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
What does it mean? I just read ACN using it also as mobile phase with H2O. I have already checked the flowcell and every part of the instrument. Moreover, when I changed solvent composition, the noise started again, after the first run in which the signal was perfectly flat. (first run: 36%ACN, 2nd run: 100%, 3rd again 36%)

(Instrument Agilent compact LC 1120, detector Uv-Vis)

Thanksss!
A "perfectly flat" baseline is *not* a good thing. There should be "some" noise (check the Agilent specifications; probably something like 5 X 10^-6 AU).

First off, are you running at a reasonable wavelength? If assume you have already attempted to zero the signal. If -0.5 AU is as close as you can get, the chances are you probably have one of the following problems:
- a severely mis-aligned flow cell
- severely solarized flow cell windows
- an occluded optical path in the flow cell (precipitated junk or material photolyzed and adsorbed onto the cell windows
- a dead lamp
- a bad electronic connection

You should be able to check the voltages from the reference and sample side photodetectors. If it's a dead lamp, then both will be zero. If it's a sample-side electrical fault, then that side will be zero but the reference side will be within spec. If it's a misaligned or occluded flow cell, then the sample side will be low (but non-zero) and the reference side will be within spec.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 43 users online :: 1 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 42 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry