lmh » Tue Oct 06, 2015 4:37 pm
well, you're not doing too badly! The two difficult peaks are actually very nearly base-line separated, which means that if you're purifying and finding that the fractions are contaminated, then the mixing-up is happening post-detector (possibly because the time-delay between the detector and the fraction collector isn't quite right, or possibly because the fraction collector is equipped with about 5m of wide-bore tubing, something manufacturers seem to do!).
Because the separation is almost good enough, you might get it to be really good enough by using a longer column, or if pressure is an issue, considering moving to Kinetex rather than Luna; I find that Kinetex is a bit less retentive, but Phenomenex can advise on the most appropriate improved column.
In terms of other changes, temperature is easy to play around with and might just make a difference; pH is useful, but only if your isomers are likely to behave differently in response to a pH change; sometimes changing solvent completely can help, and is easy to do, but it's a bit like pressing a "randomize" button: it may just move the peaks around, and they may end up in worse places than they are now.
Changing column is expensive, but if you want to avoid a difficult and really expensive thing like a chiral column, the more "rigid" reverse phase columns are reckoned to be better at separating isomers - i.e. short-chain things, which have less flexibility to wrap themselves round the analyte, and whose interactions with the analyte are therefore more dependent on the exact shape of the analyte.
By the way, does anyone know why, given modern fiber-optic technology and miniature detectors, why manufacturers don't supply a fraction collector with a miniaturised PDA-cell built into the very end of the dispensing tip? It might not give the best signal ever, but sometimes it would be really nice to know (roughly, at lower sensitivity) what's happening where it matters, rather than (with high quantitative precision) 200uL upstream.