Standby procedure for Shimadzu QP2010 Ultra
Posted: Tue Sep 29, 2015 12:04 am
Hi all.
Just a simple question about proper settings for a Shimadzu QP2010 Ultra when not in use. We only use the instrument once every few days at the moment. It is left running at all times and I don't use the shutdown procedure unless the instrument will be inactive for a week or more. What are appropriate settings for leaving the instrument in a "ready to go" state? So far I am leaving it at the same initial temperatures and pressures used in our analysis method but I am not sure how appropriate these are over the long term.
Column oven: 50 C
Injector: 250 C
Ion source: 200 C
Interface: 250 C
Carrier gas pressure: 53.5 kPa
Total flow: 10 ml - split
Column flow: 1 ml
Purge flow: 3 ml
Is it better to lower the injector and interface temperatures or is this unnecessary?
Thanks for any assistance.
Just a simple question about proper settings for a Shimadzu QP2010 Ultra when not in use. We only use the instrument once every few days at the moment. It is left running at all times and I don't use the shutdown procedure unless the instrument will be inactive for a week or more. What are appropriate settings for leaving the instrument in a "ready to go" state? So far I am leaving it at the same initial temperatures and pressures used in our analysis method but I am not sure how appropriate these are over the long term.