Quinine in tonic water using is isocratic methods
Posted: Tue Sep 22, 2015 4:08 pm
by danw9
Hi
Currently I'm trying to develop a method for quinine analysis but whatever I do I only get a 60% recovery. I only have a uv detector but I do have a column oven.
Re: Quinine in tonic water using is isocratic methods
Posted: Wed Sep 23, 2015 8:50 am
by Gerhard Kratz
I guess your column is a C18 reversed phase column. What sample preparation are you doing? It can be that you will see effects from interferences with other compounds like salicylic acid etc.!
If you get a chance to get a fluorescence detector, take it.
More details about your chromatography conditions will help.
Posted: Wed Sep 23, 2015 9:42 am
by kubowicz.tomasz
Hello
Do you need to separate quinine and quinidine?
Regards
Tomasz Kubowicz
Posted: Wed Sep 23, 2015 11:47 am
by danw9
I'm using a phenomonex C18 RP polar column, we use quinine hydrochlorite in the standard and convert that to quinine for the hplc. I make my standard to 60ppm as this is the answer im looking for on our samples but currently im getting 40ppm after converting my result to quinine sulphate. I run a degassed sample of either diet or full sugar tonic water.
Posted: Wed Sep 23, 2015 1:57 pm
by Peter Apps
Is it the standard or the samples that gives you a result of 40 ppm (presumably mg/l) ?
If it is the standard, how do you do the calibration ?
If it is the sample, how do you know that it really contains 60 ppm ?
Posted: Wed Sep 23, 2015 2:13 pm
by danw9
The standard contains 60 ppm. My samples come out at around 40ppm. They have to get sent away for analysis presently but want to bring it in house
Posted: Wed Sep 23, 2015 2:26 pm
by tom jupille
Uhh, is it possible that your samples actually contain 40 ppm?
I'm not trying to be sarcastic here. How did you actually measure the recovery (spike a matrix blank? standard addition?)?
Posted: Wed Sep 23, 2015 2:28 pm
by HPLCaddict
How do you perform the "conversion" of hydrochloride to sulfate?
Is it a mathematical conversion (i.e. are you converting the different molar weights in the calculation) or are you actually trying to "chemically" transform the salt?
If it is a mathematical conversion, I'd first check if there isn't some fault in the calibration introduced by some sort of mathematical quinine/quinine sulfate/quinine hydrochloride mixup.
Posted: Wed Sep 23, 2015 2:32 pm
by danw9
The results we get from the external samples are always in and around 60 ppm as they should be. I run a blank and get no peaks. Only thing I can see that there is something in the tonic limiting detection where as a standard of just quinine comes off as the should. Its all mathematical conversion I've checked and double checked the calculation as that's what I first thought was the issue