- Posts: 21
- Joined: Mon Aug 22, 2005 11:50 pm
Specifically, we are analyzing lavage samples that contain a wash buffer and suspended cellular matter. The sample prep itself is very simple. We centrifuge the lavage, and analyze the supernatant directly. The pellet is resuspended by treating with 3M urea and PBS and then analyzed. The final volume of both solutions is the same.
What we are finding is that upon applying the internal standard prior to centrifugation, most of the internal standard (a deuterated form of the component of interest) is in the supernatant. There is however a good amount of the component in the precipitate. Whether it is an issue of how much time the component and internal standard are in contact with the biological material, they do not seem to proportion equally between the two fractions.
Even if they did partition equally, I still argue with myself on how to calculate the concentration due to the internal standard being split between fractions. I'm thinking we would have to sum the corresponding areas in the two fractions and calculate the ratio then. Has anyone had similar experience that could offer some advice? Thanks!
Magee Women's Research Institute