Hi forum,

I'm having an issue with the Lab Solutions software in that even though I am not overloading the detector, when I look at the chromatogram, I am seeing my peak go flat (at the top of the peak) for about 20 seconds. I get the feeling I'm wrestling with a display setting I don't know how to change, especially since I've run the exact same sample with another method (same running parameters) back-to-back and don't have this problem.

Any help?

Thank you!
Just to check the obvious... are you viewing the chromatogram at an appropriate zoom level? Right click on the chromatogram and select "Initialize Zoom". Alternatively you can right click and select "Display Setting". Then tick the "Full chromatogram" checkbox and enter 100% for the area setting. That shows you the entire chromatogram regardless of how many peak groups or data acquisition periods there were in the run.
What is your detector? LCMS? or PDA?
Check your offline data and see if error persists there or not?
3 posts Page 1 of 1

Who is online

In total there are 543 users online :: 0 registered, 0 hidden and 543 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 543 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry