GC/Headspace carryover

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hello, I am very new to GC and Headspace and I'm looking for some good advice and tips.
I am following the residual solvent analysis from the EP 8.0 2.4.24.
The column I am using has not been extensively used, it has recently been re-installed and the runs following have peaks detected in my blanks.
I ran a blank (water) first followed by a reference solution and followed by another blank. Both blanks before and after the reference solution have peaks.
I have increased some headspace parameters; the transfer line, loop temperatures, vial pressure and purge time to ensure there is no condensation occurring and to push the sample through but I have found no difference.

The instruments I am using are Agilent 7697A headspace and 7890B GC
column I am using is an Agilent DB-SELECT 624UI: 30m, 320u,1.8u
Gas type He
Inlet Mode Constant Flow
Inlet Initial Time 20.00min
Inlet Initial Flow 1.500mL/min
Inlet Initial Pressure 10.0psi

Oven ramp:
Rate degC/min 10.00
Final temp 240
Hold time 20.00
Total time 60.00

Inlet:
split
140degC
split ratio 5.0
Split flow 7.5ml/min
purge flow 0.000ml/min
purge time 0.00(min)
vent pressure 0.000ml/min
vent flow 0.00
septum purge 3.0ml/min

Back Detector:
Detector Type FID
Makeup Gas Type N2
Flow Mode Constant Makeup
FID Temp Setpoint 250(Deg. C)
FID Makeup/Total Flow 30.0(mL/min)
FID LitOffset 2.0(pA)
FID Fuel Flow 30.0(mL/min)
FID Oxidizer Flow 400.0(mL/min)

Headspace set parameters:
Temperature:
Oven 80
Loop /valve 95
Transfer line 110
Time:
GC cycle time 70
Vial equib time 60
Pressure equip time 0.50
Inject time0.50
Vial:
fill at flow to pressure
fill pressure 15 psi
fill flow 50.00mL/min
loop fill ram rate 20psi
loop final pressure 10.05
loop equilibration 0.05
vial size 20mL
shaking on
Peaks in blanks don't necessarily mean carryover.
Check different water as blank.

Are the peaks from your 2nd blank higher than from 1st blank. If yes, that might be carryover.
Hello

I'd recommend bake out all heating zones in headspace sampler (loop, valves, transfer line). Put them all to their maximum allowed temperature and leave for 1 hour. If this procedure won't help I would replace sample loop.
And one more thing: for checking if headspace sampler is free from contamination run blank as empty, capped vial (do not use water)
Good luck :)

Regards

Tomasz Kubowicz
3 posts Page 1 of 1

Who is online

In total there are 173 users online :: 1 registered, 0 hidden and 172 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 172 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry