HPLC/UV with high noise and EZ Chrom software troubles!!
Posted: Wed Sep 16, 2015 2:16 pm
by AriadneY
Hello to everyone,
this is my first question here! And it could be very easy, but it is making me so annoying...
I am trying to use an Agilent LC compact 1120 for the determination of phenols in waste and water, with eluent gradient ACN/H2O, but as I start checking the baseline the noise test gives me 4000 as agains the ideal value of 50! I have cleaned the cell with H20/H2SO4, checked every part of the instrument, and nothing changed. The software EZ Chrom Elite gives me the instrument status allright.
Another issue which makes me crazy is that everytime I open a new chromatogram or I run a new sample, this is automatically opened with other old chromatograms that I cannot either delete!
I will be very grateful to anyone will answer.
Thank you!
Re: HPLC/UV with high noise and EZ Chrom software troubles!!
Posted: Thu Sep 17, 2015 6:39 am
by HPLCaddict
Hello to everyone,
First of all, check what the baseline looks like with the pump turned off. Is it still noisy? Then the problem is within the detector - maybe a new lamp is necessary? If the noise is much better with the pump off, than have a look at the pump - check-valves working properly, any leaks, pressure constant, mixer OK,...?
You mean overlaid chromatograms? Try to right-click on the chromatogram and select "clear overlays".
Posted: Thu Sep 17, 2015 10:02 am
by AriadneY
The instrument doesn't allow me to check the baseline with the pump off, it automatically turns it on.
For what it concerns the chromatograms, when I click on "clear overlays" nothing happens. Moreover, I deleted these chromatograms (it always opens the same three chr.) and now it opens just the window without any track.
I've started to use this instrument for few weeks...
Posted: Fri Sep 18, 2015 1:51 pm
by HPLCaddict
Can't you turn the pump off in the "pump" dialog WHILE the baseline check is running? That's what I usually do - let the baseline check run with the pump on for a few minutes and then turn it off, so I can see directly the change in the baseline noise.
Posted: Mon Sep 21, 2015 2:16 pm
by AriadneY
I have tried to use the Agilent Lab monitor and diagnostic software, even though I can only run some tests because it requires special kits that I cannot either find on the website. Who has used it and can tell me something more about?
I cannot turn the pump off during the run, it doesn't allow any change of the conditions during it.
Posted: Thu Sep 24, 2015 11:38 am
by uzman
Let the pump on , but do not connect it to the detector ( or column ).
Simply route the pump outlet to waste , and fill the detector cell with a syringe.