ESI Response not stable
Posted: Wed Sep 09, 2015 2:22 pm
Hi everybody
I have a very annoying issue regarding the stability of response in ESI ionization. See the data table, tells more than words:
Recovery is calculated reom a quadratic calibration curve. Injections 10 and 11 are part of the calibration curve, later injections are from the same vial.
Any idea what could have happened? If I see such a trend (between injections 10 and 40), is my calibration curve reliable? I don't think so.
This is not a one-time observation, this happens very frequently on this machine.
Details on the method: Separation of ethanolamine in 30 mM NH4formate / ACN gradient on a Primsep column. Retention time and pressure are stable. MS is a Quattro Ultima Pt operating in SIM mode, ESI+. There is a little salt buildup on the baffle plate, but in my opinion this should cause a downwards trend of sensitivity rather than an upwards trend, and especially not such a sudden drop in the middle of a sequence.
"Funny" detail: Without cleaning the source, using the same eluent and samples, the signal was back to 100 % and remained stable over a sequence of similar length.
Greetings,
Jörg
I have a very annoying issue regarding the stability of response in ESI ionization. See the data table, tells more than words:
Code: Select all
Inj. No. Response recovery
10 134101.063 97.7%
11 137028.422 100.0%
21 154454.938 114.1%
22 160899.297 119.4%
39 181004.703 136.3%
40 182717.781 137.7%
57 116918.531 84.2%
58 116183.773 83.7%
Any idea what could have happened? If I see such a trend (between injections 10 and 40), is my calibration curve reliable? I don't think so.
This is not a one-time observation, this happens very frequently on this machine.
Details on the method: Separation of ethanolamine in 30 mM NH4formate / ACN gradient on a Primsep column. Retention time and pressure are stable. MS is a Quattro Ultima Pt operating in SIM mode, ESI+. There is a little salt buildup on the baffle plate, but in my opinion this should cause a downwards trend of sensitivity rather than an upwards trend, and especially not such a sudden drop in the middle of a sequence.
"Funny" detail: Without cleaning the source, using the same eluent and samples, the signal was back to 100 % and remained stable over a sequence of similar length.
Greetings,
Jörg