Problem with a GC MS from ThermoFisher - Automatic EI tuning

Posted: Wed Sep 09, 2015 5:58 am
by KevinBe
Good morning,

I'm everything but a specialist in GC MS but recently I had a lot of problem with it...

I have many errors when I perform an autotune...

I'm attaching the image and I would be glad if anyone could help.

Thanks a lot,


Kevin

Image
Image
Image
Image

Re: Problem with a GC MS from ThermoFisher - Automatic EI tu

Posted: Sat Sep 19, 2015 11:09 am
by CHromatorgasm
Here is the manual. GTFY

http://www.thermoscientific.com/content ... W-RevB.pdf

Compased to the example tune, the intensities on this instrument are extremely low. I would start by trying to figure out why that is. Most likely dirty source or calibration compound is running out. Have you checked for leaks?