I have been having difficulty with my GC/MS (Agilent 7890A with 5975C MSD) on certain temperature program methods. On methods that begin at 100C, there is a huge peak that comes off at about 800,000 abundance and drifts downward to less than 10,000 after about 8 minutes. On methods that begin at 200C, the baseline is between 2000 and 4000. I don't know what is causing this baseline problem for the lower temperature methods. I have tried baking the column, changing the gold seal, liner and septum. I've also clipped my column, but the peak still remains. Has anyone else had this problem or know what to do?
Here are the parameters for one of the 100C method:
Inlet:
Heater - 250C
Pressure - 10.244 psi
Total flow - 54mL/min
Septum purge flow - 3mL/min
Oven:
Equilibration time: 1 min
Initial: 100C (hold 2 min)
Ramp: 20C/min until 300C (hold 18 min)
Parameters for one of the 200C methods:
Inlet:
Heater - 250C
Pressure - 16.089 psi
Total flow - 54mL/min
Septum purge flow - 3mL/min
Oven:
Equilibration time: 1 min
Initial: 200C (hold 0 min)
Ramp: 20C/min until 280C (hold 8 min)