Deactivated silica in transfer line

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hello all,

i'm thinking about putting a piece of deactivated silica in the transfer line (and couple it to my analytical column) instead of the end of the analytical column itself. The system is used to analyze PAH's and transfer line is held at 300°C-320°C to minimize PAH tailing.

The reason i want to do this is because I think that then:
1) the column itself is exposed to lower temperatures overall and should show less degradation (peaks of higher boiling PAHs start to tail after a while and trimming the transfer line piece helps)
2) I could actually use a lower transfer line temperature, since deactivated silica is far less retentive (analytes migrate at 150°C less than they would do from a coated column)
3) Replacing the piece in the transfer line is cheaper and would not impact the chromatography, since the analytical column does not have to be trimmed.

Are my assumptions correct? If so, which consideration should I take when choosing diameter, length etc... for the deactivated tubing.
The analytical column I use is a Select PAH 15m*0.15mm*0.1µm from Agilent. I'm thinking about 0.5m of 0.15mm or 0.12mm i.d. tubing.

Thanks in advance for the input!

Kind regards,

Jasn
Presuming that this is the transfer line form GC to MS (and not from a sampler to an inlet for e.g.) then it might work. With such a short narrow column any tiny dead volume in the connector will significantly degrade the separation.

Peter
Peter Apps
Peter,

Thanks for the info. It is indeed the transfer line from the GC to the MS. I know that any dead volume could degrade the chromatography significantly.

Any considerations regarding dimensions of the silica tubing to use?


Jasn
Peter,

Thanks for the info. It is indeed the transfer line from the GC to the MS. I know that any dead volume could degrade the chromatography significantly.

Any considerations regarding dimensions of the silica tubing to use?


Jasn
I don't see any reason not to use the same dimensions as the column - just add the extra length to the column in the GC setup and the software will deal with the flows and pressures.

Peter
Peter Apps
4 posts Page 1 of 1

Who is online

In total there are 33 users online :: 2 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry