I am developing an HPLC method and not in each run an unexpected peak appearing in the void volume (around 3mins). Im using a gradient method with phosphate buffer pH 3.2 in A and a mix of MetOH and ACN in B in a Hypersil CN 250x4.6 column. My diluent is H2O and when I filled 5 different vials with water then in two of them appeared this peak whilst the rest three were clean of this peak. The area of this peak varies between positive vials and when i rerun a positive vial the area of this peak remains the same. I have different HPLC systems and different columns and the problem persists. So dont seem to be a carry over or a column or a system problem. Can you help with any ideas? Is there any possibility the HPLC vials or their septon closure system to be the reason of this problem?
Thnx in advance