I´m tyring to analyse Fosfomycin Tromethamine using USP method. In PF 36(2) says:
BRIEFING
Fosfomycin Tromethamine. Because there is no existing USP monograph for this drug substance, a new monograph is being proposed adopted from the official European Pharmacopoeia 6.0 monograph. The liquid chromatographic procedure in the Assay and in the test for Organic Impurities is based on analyses performed using a Zorbax NH2 brand of L8 column. The typical retention time of fosfomycin is about 9 min.
Conditions: MP Potassium phosphate buffer, monobasic. 10.98 g/L, Flow 1.0 mL/min, detector RID, Temperature (column and optical) 35 °C. My HPLC is an Agilent 1100, and made preventive maintenance 3 months ago and replaced MGCV because it wasn´t working well.
First, I didn´t have the Zorbax column, so I used a like new Luna amino column (less tan 50 inj.). Didn´t work well, broad peaks and unresolved impurities because of tromethemine broad peak at front. Tried to clean it by phenomenex cleaning procedure, and didn´t worked. same chromatogram.
So I purchased Zorbax column. First peak came at 13 min, not at 9 min! But were nicer, thinner and peaks in front were resolved. So I changed Mp, new batch, and same problem. I made a little sequence and peak started to shift, and elutes earlier moving in each injection, always less rt, and starting to see a loss in performance, because peaks started to broad.
Well, I checked the equipment, did a pressure and leak test and didn´t pass. Called the service, and changed the outlet ball valve. The instrument then was OK, pressure and leak passed.
But the problem persists, and now the peak elutes at 7.2 min and decreasing and broadening.
Things already tried and didn´t worked:
Bypass the multi gradient valve
Changed the guard column
Using no guard column
Wash the column like agilent says (twice, first using only ACN, the second try was using DMSO too)
Now I´m getting back to Luna column and see if rt is staying the same or happens the same with Zorbax and starts to move and elute earlier tan before and never stop! Perhaps is the instrument and I´m not seeing it! And next I´m doing the QC test like the one Agilent did with the Zorbax to check performance, because I can´t believe A new column lasted less tan 50 injections.
BUT PLEASE, ANY IDEA!?!?!?!?