calculating dilution factor
Posted: Tue Aug 25, 2015 10:21 am
Hi all.
I have a question about calculating concentration of HCP in fractions collected from an AIEX run and I can't get my head around it.. i think it's a pretty simple problem but i'm not sure my logic is correct.
background info: CHO harvest feed is injected onto a 1ml AIEX column using an Akta avant. 2ml fractions are collected.
procedure:
I take my harvest and dilute it 1:4 in buffer A.
I inject 30ml into the 1ml column (at 1ml/min).
I get a wide FT flat peak, an elution peak and a tiny peak at the end (strip). (see chromatogram below)
I'm collecting 2ml fractions, so I get 15 fractions for the FT, 7 for the Eluate and 1 for the Strip.
I want to measure the amount of HCP in in the 3 peaks, so what i did and I pooled the fractions together (for each peak) and used an assay to measure HCP. The results I'm getting are too low though, and i'm afraid that i'm not taking into account the dilution correctly.
how would you do it? Do i need to multiply the result times the dilution factor? how would you calculate the dilution factors?
thank you!
https://www.dropbox.com/s/k8ttpqrwxsqdv ... 5.JPG?dl=0
procedure:
thank you!