I have an Agilent 7683b injector on my ECD instrument. It is getting a bit creaky in its old age, and I was wondering if anyone has ever swapped one for a 7683 tower, or vice versa... will it work with no issues? sit there like a lump? burst into flames? any help greatly appreciated, thanks.
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
Check document in the link below :
http://www.agilent.com/cs/library/userm ... enance.pdf
You have all compatibility information there. All you need to do is to check firmware in GC and sampler/tray and make sure it is compatible.
Regards
Tomasz Kubowicz
Check document in the link below :
http://www.agilent.com/cs/library/userm ... enance.pdf
You have all compatibility information there. All you need to do is to check firmware in GC and sampler/tray and make sure it is compatible.
Regards
Tomasz Kubowicz
-
- Posts: 9
- Joined: Thu Sep 03, 2015 12:49 pm
if you actually already have one on site you need to check its contact point.
some use a 'clicker' in the tower itself that the pole on the GC goes into.
some have a separate flange on the bottom that swivels onto the contact plate when in position.
also you need to check if it is set up with the sample sample turret (you can't swap a turret that holds 6 solvents, 4 waste, and a sample for one that holds 2 solvents, 2 waste and a sample).
also, depending on the interface with your computer you might even need to get a new card for your controller for the tower to plug into.
KDM
some use a 'clicker' in the tower itself that the pole on the GC goes into.
some have a separate flange on the bottom that swivels onto the contact plate when in position.
also you need to check if it is set up with the sample sample turret (you can't swap a turret that holds 6 solvents, 4 waste, and a sample for one that holds 2 solvents, 2 waste and a sample).
also, depending on the interface with your computer you might even need to get a new card for your controller for the tower to plug into.
KDM
3 posts
Page 1 of 1
Who is online
In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 14 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 14 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science