Small Signal Response?
Posted: Thu Aug 20, 2015 9:00 am
Hello,
I am trying to analyze the product gases of an alkaline water electrolysis plant, being hydrogen/oxygen mixtures. In this case gas concentrations of 0-1 Vol.-% H2 in (99-100%) O2 or the same amounts of O2 in H2 occur.
I found an old HP 5890 Series II with a TCD in our stock, which should do the job. As a column I am using a new HP-Plot Molesieve with Argon carrier gas. Unfortunately the signal I get seems to be very small, so that I am not really able to measure quantitatively.
The settings are the following at the moment:
Oven Temperature 35 °C
Injector Temperature 60 °C
TCD Temperature 180 °C
Column Head Pressure 18 psi
Split Vent 23 ml/min
Purge Vent 2.6 ml/min
Column Flow 3.1 ml/min
Make-Up Flow 3.0 ml/min
Reference Flow 15 ml/min
Injection Volume 1 - 2 ml
I changed all these parameters in the last weeks and injected a testgas with following composition:
0.2 Vol.-% H2
0.2 Vol.-% O2
0.5 Vol.-% N2
Rest Argon
The chromatogram I got with a injection of 1.5 ml testgas looks like that:

In this case the S/N-Ratios for the components are the following:
H2 - 29 (0.2 Vol.-%)
O2 - 5 (0.2 Vol.-%)
N2 - 8 (0.5 Vol.-%)
The signal response seems very low to me. Is there any way to increase the signal, am I injecting too much/too less or do I have to change the method? I would appreciate any hint to improve my results somehow.
Thank you very much in advance,
tehzap
I am trying to analyze the product gases of an alkaline water electrolysis plant, being hydrogen/oxygen mixtures. In this case gas concentrations of 0-1 Vol.-% H2 in (99-100%) O2 or the same amounts of O2 in H2 occur.
I found an old HP 5890 Series II with a TCD in our stock, which should do the job. As a column I am using a new HP-Plot Molesieve with Argon carrier gas. Unfortunately the signal I get seems to be very small, so that I am not really able to measure quantitatively.
The settings are the following at the moment:
Oven Temperature 35 °C
Injector Temperature 60 °C
TCD Temperature 180 °C
Column Head Pressure 18 psi
Split Vent 23 ml/min
Purge Vent 2.6 ml/min
Column Flow 3.1 ml/min
Make-Up Flow 3.0 ml/min
Reference Flow 15 ml/min
Injection Volume 1 - 2 ml
I changed all these parameters in the last weeks and injected a testgas with following composition:
0.2 Vol.-% H2
0.2 Vol.-% O2
0.5 Vol.-% N2
Rest Argon
The chromatogram I got with a injection of 1.5 ml testgas looks like that:

In this case the S/N-Ratios for the components are the following:
H2 - 29 (0.2 Vol.-%)
O2 - 5 (0.2 Vol.-%)
N2 - 8 (0.5 Vol.-%)
The signal response seems very low to me. Is there any way to increase the signal, am I injecting too much/too less or do I have to change the method? I would appreciate any hint to improve my results somehow.
Thank you very much in advance,
tehzap