Alkenes separation on a non polar column
Posted: Wed Aug 19, 2015 2:28 pm
by Edithv
I have a mixture of 1-pentene, trans-2-pentene and cis-2-pentene. is it possible to separate them on a non polar column - 100% dimethylpolysiloxane phase (30 m x 0.25 mm, 0.25 µm)?
and my second qouestion....
on the same column, what are the parameters for the separation of 1-methylcyclohexene, 3-methylcyclohexene and methylenecyclohexane?
I'll be glad for a quick answer!
Thank you,
Edith
Re: Alkenes separation on a non polar column
Posted: Wed Aug 19, 2015 11:01 pm
by rb6banjo
My gut feeling is that you're going to have a hard time with that separation on that column. It's not a very selective phase and those are some very similar materials.
Posted: Thu Aug 20, 2015 5:59 am
by Edithv
I thought so, too. I wanted to be sure about that...
Thank you for your quick reply.
Posted: Thu Aug 20, 2015 12:32 pm
by rb6banjo
You have a good chance on an alumina column:
http://www.restek.com/chromatogram/view ... 7C109-66-0
For the cyclohexane's you might have a chance. Restek doesn't have an app note on that. If you have a standard, make it up and shoot it. Just don't make it too concentrated.
Posted: Thu Aug 20, 2015 2:14 pm
by Edithv
It's for a student lab, there is no budget for a new column...
but I will try the methylcyclohexene isomers.
thanks for the advise!
Posted: Thu Aug 20, 2015 10:13 pm
by rb6banjo
A good system for the 1 phase might be ethylbenzene, p-xylene, and o-xylene. All are isomers. You can't separate m- and p-xylene from each other on this phase but you can separate EB and o-XY from m- and p-.
Posted: Fri Aug 21, 2015 9:37 am
by Edithv
I was planning a student project that compares product distribution of 1 and 2-pentanol elimination reaction.
Thank you, anyway...
Posted: Fri Aug 21, 2015 7:56 pm
by rb6banjo
Ah. I thought perhaps it was an analytical lab. The compounds I mentioned are contaminants (BTEX) that show up in ground water if a gasoline storage tank is leaking somewhere.
You might have success with the methylcyclohexanol system and the 1 phase. Generally, if the analyte is heavier, it gets easier to separate closely related compounds on that phase.
Posted: Tue Aug 25, 2015 10:22 am
by Edithv
It is easier, I have just succeeded to separate them...
Posted: Tue Aug 25, 2015 3:06 pm
by rb6banjo
Very good! Good luck with the experiment in the hands of the rookies.
Posted: Wed Sep 02, 2015 8:45 am
by Edithv
Thank you