The pressure of LC fluctuates

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I was running LC/MS. After injection, the pressure keep climbing. After 5 min maybe, the pressure started to fluctuate from between 100 and 60psi. Like climbed to 100 and back to 60psi and then climbed to 100 again with 1psi/s. The thing is when I wash the column which 100% ACN or water/ACN 50/50 , it is stable. And I did not find any leaking!

Type of LC?
Flow (gradient) program?
Mobile Phase composition(s)?
Thanks,
DR
Image

Like the Dr. ordered, provide more information. But it sure sounds more like a HPLC issue, not the detector. Without that information, I'd say to degas solvents better, purge out all/any bubbles, disconnect detector, and run pressure check tests if the pressure still fluctuates. It could be pump problem or a mixing problem, must isolate the issue. To rule out mixing issue, put all solvent leads in one reservoir, purge out all dead volume, and see if problem goes away (that way same will be pumped no matter how the individuals are mixed).
3 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry