7683 tray arm dragging.

Posted: Mon Aug 10, 2015 4:30 pm
by MSCHemist
Hi all

I am noticing my 7683 tray arm is not moving smoothly. The tray arm rotates fine and the gripper goes up and down fine and it works overall but the arm when it moves forward and especially back from the tower or whereever makes a dragging crunching noise like it is overcoming some sort of drag/friction. Any idea how I can fix it?

Re: 7683 tray arm dragging.

Posted: Tue Aug 11, 2015 12:58 pm
by dblux_
Hi all

I am noticing my 7683 tray arm is not moving smoothly. The tray arm rotates fine and the gripper goes up and down fine and it works overall but the arm when it moves forward and especially back from the tower or whereever makes a dragging crunching noise like it is overcoming some sort of drag/friction. Any idea how I can fix it?
Lubrication of the rods may help. (no need to disassembly - access from underneath)

Re: 7683 tray arm dragging.

Posted: Tue Aug 11, 2015 6:02 pm
by MSCHemist
thanks a few droplets of wd-40 rubbed arround with a cotton swab on the rods and it is moving a lot smoother with less crunching over the plastic teeth.

Re: 7683 tray arm dragging.

Posted: Fri Aug 21, 2015 1:00 pm
by gcguy
is there anything WD40 can't fix!!!
I use it on the needle guide spring on Agilent towers.