SCIEX Analyst Issue with Agilent 1100 DAD

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
Hi,

We've just added Agilent 1100 DAD (G1315A) to our LC-MS (Agilent 1100 stack, with SCIEX API 2000, all controlled by Analyst 1.4.2 software).

Everything looks great in the configuration, method set-up, etc., but when I try to run a method using the DAD, the status of the DAD changes to "Wait" in the software. The injection happens and the method runs, but no data file is created and the software stays in pre-run mode, unresponsive.

Has anyone else encountered similar issues with Analyst software controlling an Agilent DAD?

Thanks,
Ben

(Re-posted this from the LC-MS forum just in case)
Eventually figured this out - I was attempting to run as a standalone HPLC system (I had no need for MS for this particular method), and set the method not to use the MS. It turns out Analyst doesn't start any acquisition if the MS isn't set to acquire. So, I set up a dummy MS method and everything was fine.
2 posts Page 1 of 1

Who is online

In total there are 7 users online :: 1 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry