Hi,
Could you tell me, why grow up the pressure when I have mobile phase with water: water/ACN=50/50, and when I use ACN 100% the pressure is normal?
1. Aqueous and mixed aqueous-organic solvents have more dissolved air in them than do pure organic solvents - are you using a degasser or other way to remove dissolved gases before the pump module?
2. Are you purging all four channels (assuming Quaternary pump here) at least 5 minutes per channel at 5ml/minute even if you're only using 1 or 2 channels, important.
3. This relates to #1 as well: if you're using a 1050 model vacuum degasser, the hold-up PER CHANNEL is like 40 ml, so you need to purge/pump through to make sure your mobile phase reaching the pump is really what you expect.