High Random Noise in 5975c
Posted: Tue Aug 04, 2015 10:22 pm
I noticed small jagged peaks that contained every mass programmed for scanning a number of times over the last month. I had been in contact with Agilent to try and determine the cause and get ahead of it but yesterday this noise became unmanageable. The instrument tunes fine - all connections have been checked and scanning in manual tune reveals nothing odd. But when I try and do a run this mess shows up randomly throughout the chromatogram.
Ran through some diagnostics with Agilent and they are recommending changing the electron multiplier at this point but it is only tuning to 1850 Ev and otherwise seems OK via the u_phonics test. Darkcurr macro for the HED also shows nothing out of the ordinary.
I was unable to post an image but the spectra look like a flea comb you'd use on your dog showing every mass or almost every mass. The random jagged peaks range anywhere in abundance from 10000 to over 10000000.
Anyone ever see this? If so what was the problem?
Ran through some diagnostics with Agilent and they are recommending changing the electron multiplier at this point but it is only tuning to 1850 Ev and otherwise seems OK via the u_phonics test. Darkcurr macro for the HED also shows nothing out of the ordinary.
I was unable to post an image but the spectra look like a flea comb you'd use on your dog showing every mass or almost every mass. The random jagged peaks range anywhere in abundance from 10000 to over 10000000.
Anyone ever see this? If so what was the problem?