Shimadzu CBM-10A PC connection problem
Posted: Mon Aug 03, 2015 11:41 am
by dr_bahram1977
Hi
I've a Shimadzu CBM-10A controller module. Unfortunately, I don't have user manual for that, and can't connect it to PC.
CBM10-Avw has serial port and it's possible for terminal connection. but, this one has no RS232 port.
any body has this module, or knows how to connect it to PC via LAN?
Thanks in advance
bahram
bahram111@yahoo.com
Re: Shimadzu CBM-10A PC connection problem
Posted: Thu Nov 26, 2015 6:53 pm
by dr_bahram1977
Any body worked with this model? need some assistance.
Posted: Fri Nov 27, 2015 9:36 pm
by uzman
You need a communication card , and a computer with ISA slot in order to make a connection with this model.
The software required is Class LC-10 and it works under Win-98.
Also a special Ethernet cable is needed for connection.
Posted: Sat Nov 28, 2015 2:47 pm
by dr_bahram1977
merhaba uzman
Thanks for your reply.
Please let me know what is ISA card for? configuration or connection? Does it replace SCL 10? or it's just for Class LC-10 and no other software?
I'm planning to use it with LCMSsolutions.
Kind regards
Posted: Sat Nov 28, 2015 9:40 pm
by uzman
Merhaba Dr.Bahram ,
ISA card is for both configuration and connection , it does not replace SCL-10. You need both.
If you look at the backside of SCL-10 , you will see two Ethernet sockets , labeled ''in'' and ''out''.
The ISA card also have two of these sockets , so you connect ''in'' to ''out'' or vice versa.
You may connect more than one controller using these socket serial , and it is called '' daisy chain ''.
As far as I know , you can use this system only for Class LC-10 software and under Win-95 or 98.
Posted: Sun Nov 29, 2015 6:58 am
by dr_bahram1977
Merhaba uzman
Thanks for your information. If I'm right, I can't use this module for CLassVP or LCMSSolutions.
another question: is there any free software( or good price) for data acquisition from RF10 FLD bu RS232?
iyi gunler kardesim
Posted: Sun Nov 29, 2015 8:02 am
by uzman
Good morning Dr.Bahram ,
Since we are in the same country , I can help you more if you use my mail : gizemlab@superonline.com
Gorusmek dilegiyle