Anybody have a clue what is going on .
2 identical (from the outside at least) GC HP6890's - peak area responses between systems differ by 25% for various methods
Same column ,liner , carrier gas, detector parameters etc.
Even engineer confirmed both systems are fine.(robustness of methods questioned but I am not convinced)
We use peak area ratios to quantify so these are the same on both systems (due to 25% loss in int std too)
Concerned for Low level methods /volatile solvents methods we only have one GC ,,
any ideas much appreciated