I would suggest C18 column with no silanols activity. The positivly charged nitrogen is interacting with silanols groups in silica and that is why you have broad peaks.
I thought of that, this is why I chose Gemini-NX, it is very well shielded.
I would go propably with Ace SuperC18 or Hypersil BDS with 0,1% TFA in mobile phase.
I'd like to avoid TFA on my mass spec...
I did once an analysis of Benzalkonium chloride on Hypersil BDS with 0,1% TFA, first time saw almost symetric peaks with quaternary ammonium cation.
Funny, I have a method for Benzylkonium chloride on a cyano column with sodium acetate with not too bad peak shape. I will try to replace the sodium for ammonium and see what happens with this compound. Thanks for waking this well lost memory.
You didn't say what your mobile phases were, but it's unlikely that they contained any chloride salts.
Yes sorry. Mainly ammonium acetate and formate in combination with acetonitrile or methanol. Sorry I can't write down all concentrations I tried out... many.
That means that your compounds entered the column in the cloride salt form and probably exchanged chloride for some other anion during the migration through the column. Since different molecules will take part in an exchange at different times during the migration, then the result is a continuum, since the different ion pairs will differ in polarity and hence retention time. That accounts for your broad peaks.
Do you really think this effect is so dramatic for dilute solutions of about 1 - 10 mg/L?
This problem is most acute with analytes with a high charge-to-mass ratio, like yours. Solutions:
1) Make sure that the mobile phase contains the same counterion that your analyte has.
2) If that's not possible, then put a high concentration of the anion that's in the mobile phase into the sample solvent prior to injection. Hopefully the anion will compete off the anion originally associated with the analyte (here, chloride).
This sounds like a very good idea!
3) Have a lot of salt in the mobile phase. Example: In HILIC of aminoglycoside antibiotics (which can have 5 amine residues in 4 sugar residues), one doesn't get symmetrical peaks unless the mobile phase contains about 120 mM salt. It can be a volatile salt, like ammonium acetate. For that matter, why not try those conditions with your compounds here? You can use our PolyHYDROXYETHYL A material, but I caution you that best results were obtained with the 1000-A pore version. That's not because aminoglycosides are so large that they posse steric hindrance problems. Rather, it's because the 1000-A material has about 25% of the surface area of the 300-A pore material, making it easier to elute these extraordinarily hydrophilic compounds.
Andy Alpert
PolyLC Inc.
This column material sounds interesting, but I don't think that our boss will invest in new column materials right now.
Anyway thanks a lot for valuable suggestions that might help with available materials (at least available to me).
Jörg