Superdex 200 Increase column blocked from one direction
Posted: Mon Jul 20, 2015 4:13 pm
by Janis Bogans
My Superdex 200 Increase Tricorn 10/300 column is blocked from right direction (from adapter to bottom), but is clear from bottom to top. What is the reason? I change top filter, but that not solved problem.
Re: Superdex 200 Increase column blocked from one direction
Posted: Tue Jul 21, 2015 12:26 pm
by danko
It doesn't sound logical.
It might be the connection - fitting problem.
Try and change the capillary and fitting.
Posted: Tue Jul 21, 2015 12:51 pm
by Janis Bogans
With other column all is ok. I think may be one of too reasons why column run in only one direction.
1) Superdex 200 is soft medium and is clogged at top or
2) something small is collected at bottom filter and clogged it.
Posted: Tue Jul 21, 2015 2:12 pm
by danko
If the column blocked inside the pressure will rise regardles of the flow direction.
If it's a fitting problem it migth occur only in a specific connection and that goes for a different column as well.
Posted: Thu Aug 06, 2015 2:05 pm
by Janis Bogans
I solved the problem asking for advice from GE Healthcare tech support, but answer is confidential. If you have problems with GE Healthcare products ask assistance from GE Healthcare tech support team.