Can anybody shed a light on what's going on?
I really appreciate any help and suggestions
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
Try tightening the four screws on each end of the analyzer, they hold the end caps on and there is a seal under them I think. Also can have a leak at the ion gauge fitting or where the calibration gas valve attaches. Another place these can leak is where the transfer line heater attaches to the side of the analyzer. We also had one that the inlet cracked right behind the threads where the column interface nut attaches, the interface finally broke right behind the threads is how we found that leak.It turned out to be a leak in detector (column is removed). I lubricated side-plate o-ring and vent valve with grease, checked them as well as GC/MSD (capped) interface. All of them didn't show a leak with argon balloon. But overall leak remains in a system an it increases with time!