GC/C/IRMS Combustion reactor - no peak
Posted: Fri Jul 10, 2015 6:29 am
Hello,
I have a problem with our combustion reactor in the lab. Since May he hadn't worked, but in this time it was fine. This week when I run a standart sample I don't see a peaks from the components of the sample. I checked with stream of Ar the connections to the combustion reactor to see if he is broken. Everything is ok. I run a re-oxidation for 2 hours. Also I injected 3 ul air to check for conductivity - again it seems that everything is fine. Then I injected 3 ul breath (5% CO2) - again I have detected a peak. I have changed the septum and inlet with new one. But I still don't see a peak. Could you give me some advices what else I can do and where may be is the problem?
Thank you in advance
I have a problem with our combustion reactor in the lab. Since May he hadn't worked, but in this time it was fine. This week when I run a standart sample I don't see a peaks from the components of the sample. I checked with stream of Ar the connections to the combustion reactor to see if he is broken. Everything is ok. I run a re-oxidation for 2 hours. Also I injected 3 ul air to check for conductivity - again it seems that everything is fine. Then I injected 3 ul breath (5% CO2) - again I have detected a peak. I have changed the septum and inlet with new one. But I still don't see a peak. Could you give me some advices what else I can do and where may be is the problem?
Thank you in advance