This is in some ways a general question on reverse phase HPLC gradients of a type I'd seen from time to time. The gradient goes something like this:

Time %A %B %C
0 min 2 20 78
10 min 20 20 60


Where A is a buffer -- 35 mM ammonium acetate, B is acetonitrile, and C is HPLC water. The column is a typical C18, and the first analyte to come off is potassium sorbate, and the second is phenoxyethanol, the second one being a more non-polar molecule of these two typical preservatives.

My question is: how does this work? The solvent concentration is the same. The water concentration is essentially the same. There's greater salt concentration, but how does that drive a reverse phase separation? Diminished water activity because of greater ionic strength? What problem does methods such as these fix? And what sort of pitfalls are there in making methods such as these?
How does this work? Elementary Dr. Watson ;-)

It’s just not a gradient elution – Reversed Phase Chromatography wise.
Rather it is isocratic elution. The buffer could have other roles but it’s not the question in this context.

Best Regards
Learn Innovate and Share

Dancho Dikov
Hello

I've never seen gradient method with buffer:water

Tom
Thanks, for a while, I was suspecting that this was hydrophilic interaction chromatography. But I don't see an analyte that needs it. Maybe another common ingredient needs it, or the developer just did it for no reason. We'll see if we can look into it at some time.
This is in some ways a general question on reverse phase HPLC gradients of a type I'd seen from time to time. The gradient goes something like this:

Time %A %B %C
0 min 2 20 78
10 min 20 20 60

Where A is a buffer -- 35 mM ammonium acetate, B is acetonitrile, and C is HPLC water.
Are you sure you got this right? It would make sense if A: acetonitrile, B: buffer, C: water. This is something that can be seen sometimes - the sense behind it is to hold the buffer concentration constant while running the gradient.
5 posts Page 1 of 1

Who is online

In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Semrush [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry