HP5890 with chemstation

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hello,
The PC that controlled a HP-5890 series II with a FID and a TCD detector doesn’t work so we buy other with Windows XP. We install the HP-IB card 82341, Agilent libraries version L.02.01.00 and a ChemStation Rev. A 10.02. We configured the interface:
Visa Name: GPIB0
SICL Name: hp82341
Logical unit: 7
Bus Address: 30
With the HP-Configurator editor I’m trying to put the GPIB address but the problem is that I’m not sure what it is… I try with some but I cannot connect the GC with the ChemStation.
Someone could help me? Thank you in advance.
isa39,

How about just setting your GC to 15?

<---
<---
<---
<---
--->
<---
<---
<---

on your card inside the GC.

Best regards,

AICMM
Agree with AICMM, use GPIB address 15. This is what my 5890/XP setup uses.
Thank you very much. It seems to work well. :)
4 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry