Hello,
The PC that controlled a HP-5890 series II with a FID and a TCD detector doesn’t work so we buy other with Windows XP. We install the HP-IB card 82341, Agilent libraries version L.02.01.00 and a ChemStation Rev. A 10.02. We configured the interface:
Visa Name: GPIB0
SICL Name: hp82341
Logical unit: 7
Bus Address: 30
With the HP-Configurator editor I’m trying to put the GPIB address but the problem is that I’m not sure what it is… I try with some but I cannot connect the GC with the ChemStation.
Someone could help me? Thank you in advance.

Re: HP5890 with chemstation

Posted: Mon Jul 06, 2015 5:37 pm
by AICMM
isa39,

How about just setting your GC to 15?

<---
<---
<---
<---
--->
<---
<---
<---

on your card inside the GC.

Best regards,

AICMM

Re: HP5890 with chemstation

Posted: Tue Jul 07, 2015 12:15 am
by aidnai
Agree with AICMM, use GPIB address 15. This is what my 5890/XP setup uses.

Re: HP5890 with chemstation

Posted: Wed Jul 08, 2015 8:59 am
by isa39
Thank you very much. It seems to work well. :)