Hi All,

Wondering if anyone has experienced similar problem.
Recently I have been having a problem with the signal stability in Quattro Micro API, I am using an ESI probe and it was working fine. Suddenly while I was trying to run direct infusion, the signal seems redundant and it keeps fluctuating badly.
I have tried to change the source parts but it is not working.

Any idea where the problem might be?

Thanks
Please let us know if you've baseline problem, or this is about your peaks.
Please let us know if you've baseline problem, or this is about your peaks.
The analyte peak appears and disappears when I am operating at low flow rate 10 ul/min, the baseline seems fine however!
Main reason for an unstable signal in infusion is the capilairy. Try to rebuild the probe and check the grafite ferrule. Maybe this one is set to loose or maybe the ferrule is damaged by overtighting. Also deadvolume on this connection might influence the signal.
4 posts Page 1 of 1

Who is online

In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry