7890A GC - 5975C MSD Tune action stopped: Gain and offset =0

Posted: Tue Jun 30, 2015 4:40 pm
by MsMarilynCurie
Hi,

I just installed and conditioned a new column for the GC-MS. Once I connected the GC to the detector (MSD) and pumped it down, I tried to tune it and it gives me this error : Tune action stopped: Gain and offset = 0 and peaks are too narrow.
The MSD does not complete tuning and the search for peaks looks like noise.
It is my first time changing a column and I was wondering if anyone had experienced something similar.

Looking forward to hearing from you guys!

Thanks,
Ms.Curie

Re: 7890A GC - 5975C MSD Tune action stopped: Gain and offse

Posted: Tue Jun 30, 2015 6:25 pm
by James_Ball
Did you open the vacuum chamber to position the column or did you use the tool to set the column length inside the analyzer.

If you got the column too far into the source it could cause a problem like this.

Also what is your vacuum reading, if you have the ion gauge installed? It should be in the 10^-5 torr or lower.

Re: 7890A GC - 5975C MSD Tune action stopped: Gain and offse

Posted: Tue Jun 30, 2015 6:50 pm
by DReggio
This happened to me once, the entrance lens had become disconnected during pump down.

Re: 7890A GC - 5975C MSD Tune action stopped: Gain and offse

Posted: Wed Jul 01, 2015 12:41 pm
by Bigbear
Try a different tune, Agilent tune files can become corrupt.