Sodium Dodecyl Sulfate interaction with PFP column

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi,

I am running a method using a polyfluorophenyl LC Column. I am analyzing Taxol samples prepared in Sodium Dodecyl Sulfate in pH 4.0 buffer solution. I am having continuous issues with inconsistent retention time and peak response reproducibility for the taxol peak using multiple columns of the same type and 57:43 Water:ACN as mobile phase. I've tried washing the column with a 50:50 Methanol Water mix but the issues keep returning. Has anyone experience using SDS on a Phenyl column, and knows of any issues there may be when using this combination?
Thanks for any advice.
Check out this source: www.chem.agilent.com/Library/.../5988-7973EN.p...
Or this source: https://www.phenomenex.com/.../HPLCDetail/Curosil
Or this source: Chromegabond™ PFP/T HPLC Columns - ES Industries
Or this source: www.sigmaaldrich.com/catalog/product/.../g005890
and many others. SDS should not make any trouble on a Phenyl column. Look for the applications to see what they did.
At what temperature you are running your method? Did you check your flow rate at each channel?
Good luck
Gerhard Kratz, Kratz_Gerhard@web.de
Check out this source: http://www.chem.agilent.com/Library/.../5988-7973EN.p...
Or this source: https://www.phenomenex.com/.../HPLCDetail/Curosil
Or this source: Chromegabond™ PFP/T HPLC Columns - ES Industries
Or this source: http://www.sigmaaldrich.com/catalog/product/.../g005890
and many others. SDS should not make any trouble on a Phenyl column. Look for the applications to see what they did.
At what temperature you are running your method? Did you check your flow rate at each channel?
Good luck
Thanks for the links but they seem to be down.
Sample temp is 23 C. Column heater 50 C. Flow rate is monitored and is seen to be constant at 1.5ml/min.
http://derpharmachemica.com/vol2-iss2/D ... 09-115.pdf
Hope that link will work.
So far I remember such problems as monitored by you lead in the past to special Taxol columns from different manufacturers.
Maybe 50°C for the column is a little bit too high, or better to say the temperature difference between sample and column too high in my opinion.
Column equilibration seems to be critical with these Taxol methods.
Gerhard Kratz, Kratz_Gerhard@web.de
4 posts Page 1 of 1

Who is online

In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry