Hi,

I am doing aflatoxin analysis by Agilent 1260,FLD detector with post column derivatization.
Conditions are
Mobile phase: 65:35 Water65:35 Meoh
Column: Poroshell 120, 50mm X 4.6, 2.7 u.
Temperature: 30 Deg C
Standard: Aflatoxin Mix fro Supelco- CRM46303

In the COA the purity is mentioned is done by UV detection at 360 nm.
How i can use this standard for quantification of Aflatoxins.
The peaks also coming different what they mentioned in the COA.

Can anybody please help on this.
The system Supelco used to test the purity of the standard is probably not the same as yours so you aren't going to get the same retention.

You can prepare appropriate calibration standards by diluting the aflatoxin mix with solvent/initial mobile phase or in your sample matrix and them along with your samples.
2 posts Page 1 of 1

Who is online

In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry