- Posts: 13
- Joined: Thu Feb 26, 2015 5:35 am
I thought I would advise that I had a error code come up on an older waters detector. The code 402, going by the paperwork tells either the lamp is failing (not enough UV light) or the cell is dry. I was sold this unit and told the lamp would need replacing someday soon as it had seen a lot of hours.
I couldn't wait for delivery of a new one so I removed the top off chassis, removed the cover box screws and lifted it out to check the glass lens.
The lens had a solar burn in it. This was clouding the lens to the point that I figured it would not let light through well.
To fix this I removed the lens, and since the burn was offset, not in centre of lens, I spun the lens around so that the burn was opposite its old position.
Then the UV light would pass through the metal slot and enter clear glass.
Upon start up and calibration. It began to work. No error message, and I could see the wavelength read out.
I have noticed however, that the absorbance value ( above wavelength on screen) drops further and further into the negatives over time.
If I hit the Auto zero button it resets back to zero, but over time, it will drop again.
I don't expect this is because the UV light strength is wearing away that fast but it may be.
Is there any other reason for the absorbance to fall into negative numbers over time without running any solvent through the cell? I can check this when the new lamp arrives to see if this is usual.